Yahoo Answers is shutting down on May 4th, 2021 (Eastern Time) and beginning April 20th, 2021 (Eastern Time) the Yahoo Answers website will be in read-only mode. There will be no changes to other Yahoo properties or services, or your Yahoo account. You can find more information about the Yahoo Answers shutdown and how to download your data on this help page.

for spectrophotometry if the absorbance change before and after a reaction is 0.001 would it still be accurate

doing an experiment with DCPIP for quantitation of vitamin C in orange juice

If possible, provide references

2 Answers

Relevance
  • 1 decade ago
    Favorite Answer

    It depends upon the S/N ratio of your instrument and the dynamic range. With a good instrument an amateur can measure changes on the order of 1E-5 au, so your signal of 1e-3 au is still large and credible. If at all possible, do the chemistry in the spectrometer so that the alignment of your sample cell is constant. To establish the validity of your result, measure the root mean square noise in the spectral region of your measurement and compare this to your signal. If the RMS noise is much smaller than your signal then the measurement is valid. (there are more sophisticated methods to extract information from noisy spectra that are beyond the scope of this forum). Next, repeat the experiment at least three times to establish that the result is reproducible.